Hello, I'm currently working on a project using Keysight E5061B with impedance analysis option. The current instrument library doesn't have option to read the impedance from the VNA. Is there a way to read the impedance values from the VNA at a particular frequency?
Solved! Go to Solution.
08-11-2020 09:34 AM
I've encountered incomplete drivers from companies before... Sometimes companies simply provide a set of common functions and not a complete set. Although you may want to check their most recent drivers just to be certain.
The programming manual can be found at.
http://literature.cdn.keysight.com/litweb/pdf/E5061-90042.pdf
Chapter 6 appears to contain what you want.
08-19-2020 02:24 PM
Thanks for the link.
Now I'm able to set impedance analysis.
I’m setting impedance analysis in my VNA with Port 1-2 Series.
SCPI.SENS1.Z.METH TSER
Then I’m selecting |Z| parameter using below command.
SCPI.CALC1.SEL.ZPAR.DEF Z
When I enquire using SCPI Command it shows the above 2 is correctly set. But I encounter another problem.
When I set both the command above the screen on the VNA doesn’t change to show we are in this mode.
The reason I require the screen to change is because I’m reading the value using the marker.
If I’m doing measurement using S11, S21, S12 & S22. The screen able to change accordingly so I’m able to read the values from the marker. But it doesn't happen when I set impedance analysis.
How can I activate the VNA window to show impedance analysis using SCPI command.