Data Acquisition with Honeywell Load Cell 31
Hi all!! I am currently trying to acquire load data with a Honeywell 31 load cell connected to the NI 9237. The samples are then displayed as an average.
While the code has no errors, the averages are not correct - a higher load returns a negative value or one smaller than when there is no load. The VI is attached, and I would really appreciate some pointers.
Also, as this is my first time connecting the 9237 and the Honeywell load cell via a NI 9949 module, I just want to double-check that my connections are right:
Honeywell    ---->    NI Pin #
+ output                 2
-  output                 3
+ excitation            6
- excitation             7
Thanks a bunch!!!
Message 1 of 10
(6,346 Views)
I verified the connection diagram with the connection you did and those seem right.
However, if this is a full bridge sensor keep in mind that the you must connect the AI+ and AI–, EX+ and EX–, and RS+ and RS– signals as  listed in the manual of the NI-9237
Luis A.
Message 2 of 10
(6,334 Views)
Thank you for your verification. However, the Honeywell Model 31 load cell has only 4 wire connections, and I use a RJ-50. In addition, could you please take a look at my VI for data acquisition and averaging? The data collected seems to be off.
Thank you so much for your time.
Message 3 of 10
(6,325 Views)
Hi rubberduckiness,
How "off" is the data? How does it differ from what you expected?
Are you using TEDS? Your VI seems to, but you said the load cell has only 4 wire connections.
There's a shipping example that demonstrates the load cell support that was added in DAQmx 9.1. I think that would be a better starting point, since it scales to force/weight units: https://decibel.ni.com/content/docs/DOC-12259
Brad
Brad Keryan
NI R&D
Message 4 of 10
(6,321 Views)
Hi, Brad. I thougt I was using TEDS (or at least that's what I ordered from Honeywell), with the load cell connected to the 10-pin NI 9949, which connects in turn to the 9237 via a RJ-50 wire. Was there a component I missed?
With respect to data, what I mean is that the average values with applied loads are not in a linear fashion. In fact, sometimes they are negative, even when I had the offset nulling on. The load cell output an expected response to increased load when I test it in the Measure and Automation Explorer, but somehow the VI makes the data screwy.
Thank you for your help.
 - Katherine
Message 5 of 10
(6,317 Views)
Katherine,
I took a look at the Honeywell 31 data sheet. If you look on page 3, there is a table called option codes. If you got option 53t, then it would be a TEDS device. Not all Honeywell 31s are TEDS. Since yours has 4 wires, you most likely didn't get a TEDS device. However, it looks like if you have one of these it will allow you to use your device with the NI 9237 as a TEDS.
The reason your code may not be working correctly is because it is configured for a TEDS device, and your sensor isn't. Once you obtain the part mentioned above from Honeywell, it may start working. Let us know!
Katie
Katie Collette
Message 6 of 10
(6,283 Views)
Hello, Katie. Thanks for the response. While my Model 31 transducer did not have a TEDS, I also purchased the NI Cable assembly with screw-terminal to RJ-50 with mounted inline TEDS on the cable. Should that have worked?
 - Katherine
Message 7 of 10
(6,271 Views)
Katherine,
I can't seem to find the cable you are describing. Do you have a part number for it?
Thanks!
Katie
Katie Collette
Message 8 of 10
(6,252 Views)
Hello, Katie. My original order form is attached. I circled both the model 31 load cell I bought and the TEDS cable.
Thanks!
Best, Katherine
Message 9 of 10
(6,238 Views)
Katherine,
My apologies--I misunderstood; I thought it was an NI cable as in purchased from NI. I found the part on the Honeywell site, and when you click data sheet it is the same link I posted earlier. Honeywell advertises this as a way to allow you to use your device with the NI 9237 as a TEDS--so in short, yes, this should work fine!
Katie
Katie Collette
Message 10 of 10
(6,230 Views)
