When I get completely stuck with GPIB I usually try to make it work with
IBIC first. IBIC is the GPIB interactive control . If you have
Measurement and Automation Explorer you can get to this through that,
otherwise, try to find and run "ibic.exe". There is too much to explain
with regard to how to use it - you'll have to study the help files (type
help at the command prompt). But here's what I might do just to see if
communication is possible at all:
ibfind gpib0 // initialize gpib board
ibdev 0 2 0 7 1 0 // you may choose different params here - check the
help
ibonl 1 // place the device on line
ibln // if this returns TRUE then your gpib card
is aware of the device on the bus
ibwrt "id?"
// or *IDN - check your device's manual for
instruction syntax
ibrd 200 // read up to 200 characters - this should
return the device's response
q // quit
good luck, Vinny
Andreas wrote:
> I've got some troubles with my HP 34401A Multimeter.
> I can't get any connection to the meter. The address is matching, the
> GPIB-cable is ok. I can't find any errors.
> The properties in the VI do not make any changes in the meter.
> What can I do?
>
> --
> Best regards
> Andreas
>
> Sent via Deja.com
> http://www.deja.com/