It doesn't find the instrument at all.
Mark
07-25-2006 10:27 AM
TO "find the instrument" the device must "respond to the *IDN? query "?
You can not expect pre-488.2 device to show up in MAX.
Ben
07-25-2006 11:14 AM
Ben,
I hate to disagree with you but the initial 'find' of an instrument has nothing to do with *IDN? or any other command string that is sent. There is a great explanation by GPIB Guru in the post at http://forums.ni.com/ni/board/message?board.id=140&message.id=17530&view=by_date_ascending&page=1 that explains what is done before MAX sends the *IDN?.
wash,
First, what is the make and model of instrument? Have you checked things with a different cable? Can you check it with a different instrument? How is the address set on the instrument? Is it with dip switches or a front panel option. Verify that the address is unique and not 0. Also, some instruments require you to explicitly turn on GPIB communication from some front panel menu.
07-25-2006 11:18 AM
"I hate to disagree with you but..."
Please do no hesitate to disagree with me.
Pushing my knowledge envelope to the limit often yeilds new.
I stand corrected!
Thank you,
Ben
01-26-2011 11:19 PM
Hello,
In my application I am conducting a pressure test, Here am using DMA Make Air Data Test Set witch I am interfacing to the PXI system through GPIB.
In MAX the particular GPIB address is detected but if i write a query "*IDN?" it did nt respond.
To debug this issue I used NI 488.2 Interactive control when i started communicating with the instrument it throws error such as EAB0 and status as TIME0 and CMPL
Here the Instrument DMA make MPS31C is 1EEE 488.1 Compliant.and NI 488.2 Ver 2.52
Please find the attached operation manual and Device communication manual of the Air data test set
Please suggest be a solution for this issue.
Waiting for a quick reply
Regards
Lingaraj
01-26-2011 11:50 PM
Did you see the *IDN? command any place in the manual? I didn't. If the vendor did not implement it (not uncommon in old instruments), just ignore the warning in MAX.
01-27-2011 12:00 AM
its not responding to any commands tat is being given in the device communication Manual.
01-27-2011 12:05 AM
That's a different problem than what you described.
I would suggest you try different termination characters. You can set it in MAX or append CR, LF, or both in the strings you send.