I'm using the Agilent N6700 LabVIEW drivers for configuring the instrument. I am able to configure the first 3 channels whithout any problem but when I try to configure the Channel 4, it throws and error -222 " Data Out Of Range". I'm able to configure the channel 4 physcally on the instrument but not from LabVIEW.
Has anyone faced this issue??
Any help would be appreciated.
regards,
Nitz
Solved! Go to Solution.
04-30-2014 03:06 AM
Hey Guys,
I kind of figured out the issue myself. The Channel 4 was grouped with Channel 3 in which case, only the lower channel (ch3) is configurable which will inturn configure the entire group. I was missing this point which I found in the data sheet.
Thank you all who looked into this post and tried to help.