Acqiris digitizer in LabVIEW, with Keysight VISA installation

I need to program a Acqiris digitizer in LabVIEW. I got their latest driver, and it requires that some Keysight libraries be installed first. So I started installing those, and they indicate that they require that Keysight VISA be installed. It said I already had VISA installed (NI, of course) and so Keysight VISA would be installed as secondary. I've had weird issues in the past when multiple VISAs are installed, so I'm hesitant to pollute my main development machine with this. I could rev up a VM, but there are complications there that I won't detail here.

The big question is: Does anyone have experience with Acqiris Drivers with Keysight VISA installed as secondary? Is this safe to do?

Thanks,

      Dave

-------------------------------------------------------------
David Thomson Original Code Consulting
www.originalcode.com
National Instruments Alliance Program Member
Certified LabVIEW Architect
Certified Embedded Systems Developer
-------------------------------------------------------------
There are 10 kinds of people: those who understand binary, and those who don't.
0 Kudos
Message 1 of 4
(1,205 Views)

A more general question that more people can potentially answer would be:

Have you installed Keysite VISA secondary to NI VISA without any problems (regardless of what instruments you were trying to talk to)?

-------------------------------------------------------------
David Thomson Original Code Consulting
www.originalcode.com
National Instruments Alliance Program Member
Certified LabVIEW Architect
Certified Embedded Systems Developer
-------------------------------------------------------------
There are 10 kinds of people: those who understand binary, and those who don't.
0 Kudos
Message 2 of 4
(1,179 Views)

No experience with that exact instrument, but we have been using HP/Agilent/Keysite instruments and some of their software along side LabVIEW since they were owned by HP.

 

I have installed Agilent VISA as secondary VISA and not had any issues with LabVIEW or Agilent/Keysite proprietary software that require VISA.

========================
=== Engineer Ambiguously ===
========================
Message 3 of 4
(1,166 Views)

Excellent. Thanks for the info!

-------------------------------------------------------------