Accuracy issue while using PCI-5102 with 1/10x probe
I am planning to use NI PCI-5102 , 15Mhz, 20 MS/s, 8 bit digitizer
The device under test is a sensor which generates pulses which needs to be measured , The following parameters are measured
1. Frequency of the pulse in range of 0 to 300 Hz with a resolution of 0.1 Hz
2. Rise time 30 us, Fall time 10us
3 High level and low level of the pulse which will be in the order of  Vcc- 0.1V
Since input range of PCI -5102 is 100mV to  10V  pp , I am planning to use a attenuator 1/10x probe
 because my Vcc is in the order of 16V DC
Now the question is will there be any inaccuracies in the measurement because of the use of attenuator 1/10x probe.?
(2,802 Views)
If you don't care about measuring the DC offset, you could use the AC coupling mode of the 5102 and a much smaller input range that would give you more resolution. With the attenuator giving you a .01V pulse, triggering might be a problem as well. If the offset is something you need, could you have the 10x probe on one channel and the AC coupled channel connected to the same point?
(2,791 Views)
Hello ,
Thanks for the your comment, I will just think over this.
(2,781 Views)
