06-20-2014 01:09 PM - edited 06-20-2014 01:11 PM
Hi all,
I am using Keithley 2440 and I am going to charge a capacitor and then measure the voltage of that capacitor to see how fast it discharges BY ITSELF. So I need to set the output to High Impedance and measure the voltage.
But my problem is how I can measure while I set the output to High Impedance. I made a vi to do that. But during discharging, Keithley voltage remains the same as charging. So, obviously I am not using High Impedance correctly.
Is there any tutorial for measuring voltage and/or current at High Impedance output setting?
06-20-2014 02:10 PM
What do you consider high impedance? You have to have the voltmeter connected in order to read the voltage. There is no way around that. Most voltmeters have an impedance of 1MOhm or 10MOhm. For most applications, that is considered open enough.
06-20-2014 04:21 PM
Thanks for the reply.
I think ~MOhm is ok for me. But the problem is that even if I connect to a resistance for example(!) and set voltage (to say 1V) and measure voltage/current for a while, and then set High Impedance output, Keithley still set output voltage to 1V and I measure 0 current/1V voltage. So there is something wrong here for sure.
06-20-2014 04:48 PM
06-21-2014 09:37 PM
Yes. I looked at the examples that come from Keithley 24XX deriver (vi attached). I suppose that impedance of voltage measurement is high enough so that it does not affect the capacitor that I am going to do measurement on. So I can use Keithley to just measure the voltage.
But, the problem with all "measurement" examples that I've seen is that you need to set an output level to be able to measure it! Why should I need to set output if I want to measure it?
Thanks!
06-21-2014 09:39 PM
Thanks!
Don't you think READING the manual and understanding how the instrument works should be your next step?
06-21-2014 09:57 PM