lecroy oscilloscope automatic acquisition

Hi, I am making a Labview program to control my Lecroy  Waverunner 374 to acquire signal automatically (for example, per 1 minute) from two channel (spescifically trace A and Trace B) simultaneously. I have a problem, where the result of the signal I got, it's not a smooth signal (It interupt with a certain noise). I tried to acquire the signal manually with ScopeExplorer and I got a good result, so the problem must be on my program. Can anyone help me with this? Thanks!

0 Kudos
Message 1 of 12
(7,883 Views)

Do not run your commands the the scope in parallel.  You are likely just confusing the crap out of it.  That VISA wire should not fork.


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
Message 2 of 12
(7,874 Views)

so, what if I want to acquire the trace from two channel simultaneously? I need to make a parallel measurement at once.

0 Kudos
Message 3 of 12
(7,872 Views)

You need to setup your scope to trigger once, let it trigger, then get each waveform.  You can then trigger again when you want the next set of waveforms.


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
Message 4 of 12
(7,869 Views)

Any other solution I can implement to my Labview code?

0 Kudos
Message 5 of 12
(7,867 Views)

Your problem is that you only have one bus to talk through.  Therefore you can only get one of the waveforms at a time.  And if you have the scope continuously sampling, the signals will not be synchronized.  That's why you need to make the scope stop sampling before you get each waveform.

 

Depending on what frequency bandwidth you are trying to capture, you might be better off with a DAQ from NI that can constantly sample multiple channels and give you both channel's data with a single call to DAQmx.


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
Message 6 of 12
(7,861 Views)

I am not sure how to use DAQ, since I am very new to Labview. I tried to run the program, it actually works, but, somehow, there is a sudden surge I got in the mid of the measurement. I wonder how other Lecroy user may solve this problem if there's any

0 Kudos
Message 7 of 12
(7,856 Views)

Hello,

 

It appears that "crossrulz" has resolved this with his recommendation to synchronize the acquisition.  You must trigger the scope one time, then go get all of your waveforms.  Then you can arm the trigger for the next acquisition.

 

The issue you appear to be having is becuase the scope is constantly triggering so as it fills up the memory with the data, you are asking for the data and it sends what it has, or you have asked for the data before it is available.

 

i.e. if you acquire 10 seconds of data, you must wait for 10 seconds for the scope to acquire it before you ask for the data. (It is actually a little longer than this because the scope must acquire the data, process any math or measurements you have setup, and then render the display for everything).

 

Use the WAIT command to synchronize your acquisitions. (There is a section in the remote control manual on page 43 that describes how to "Transfer Waveforms at High Speed").

 

Here is a link:  Remote Control Manual

 

Regards,

 

Leonard Brown

Applications Engineer

TeledyneLeCroy

Message 8 of 12
(7,839 Views)

ok, I will modify my program a bit then, thanks for the suggestion. I will let know if I can do it correctly as advised and if it finally works 🙂

 

Thanks for the replies, friends.

0 Kudos
Message 9 of 12
(7,820 Views)

Dear all,

 

I try to modify my code with adding the trigger. However, it goes totally wrong, where the setting of the oscilloscope was reset and my program won't even running (I attach the new code with trigger: run07mod totally wrong.vi)

 

So, I will review my vision of my program:

 

My program:

lala.png

 

where I can determine the name of the files, aquisition interval and number of aquisition.

(I have the .VI attach at my first comment)

 

I'd like to make a program where it:

1. Automatically make a number of data aquisition after waiting for a specified time

2. Simoultaneously take 2 traces from the oscilloscope after the sweep complete But sometimes, I need to specify the program to wait for several minutes/hours before make a new aquisition

3. Save the file in .trc format.

4. Saqve the file in .txt format

 

FOR THE run7(ok).vi (BEFORE I ADD TRIGGER), what happened is:

 

So far it can save the data in .txt format, but the .trc goes wrong even though I am sure the sweeping time is always less than the waiting time (aquisition time). So, there was no technical mistake like that.

 

The trace is suscessfully acquire, but, it was wrong!

What I get:

salah1.png

 

for trace A and 

 

salah2.png

 

 

for trace B.

 

It should be like this:

yaya.png

 

for trace A and

 

yaya2.png

 

for trace B.

 

Can anybody suggest me what's gone wrong here?

 

Thank you!