You do not need to install Agilent VISA in order to use your Agilent GPIB adapter with NI-VISA. The followign KnowledgeBase explains how to configure NI-VISA to see your Agilent adapter using the Tulip passport:
KnowledgeBase 20KG1C7Z: Can I Use Both National Instrument and Agilent Controllers in the Same Syste...
The later versions of the Agilent I/O Libraries replace one of the NI-488.2 DLLs, creating the conflict you described in your post. Remove the Agilent I/O Libraries and Aglient VISA, and you will be able to communicate with both controllers through NI-VISA following the instructions in the above KB.
Jason S.
Applications Engineer
National Instruments
03-08-2006 02:51 PM
03-08-2006 03:27 PM
03-08-2006
03:43 PM
- last edited on
04-06-2026
04:57 PM
by
Content Cleaner
Which version of the Agilent IO Libraries are you installing? Version 14.1 is incompatible with NI-488.2, in which case you will need to install a prior version of the Agilent I/O Libraries.
The following KnowledgeBase describes this problem in more detail:
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019L7oSAE&l=en-US
Jason S.
Applications Engineer
National Instruments