10-15-2018 07:07 AM
I'm trying to connect to a Yokogawa GM10 DAQ Device with LabView. I've been unable to get any sort of communication through USB(COM), or ethernet. The device will not register in NI-MAX. The drivers list the "GM10" as being compatible.
Has anyone been successful in communicating with the GM10?
Solved! Go to Solution.
10-15-2018 02:39 PM
First, did you follow the Yokogawa GM10 Instrument Configuration Manual before trying to connect to it?
10-16-2018 09:38 AM
Does the device appear in your Device Manager?
David F.
Applications Engineering
National Instruments
10-16-2018 02:47 PM
Not sure where it is in the manual, but got it working.
In NI-Max, the LAN resource has to be added as a "Manual Entry of Raw Socket." The IP address should be entered as normal, but the Port Number must be "34434". The Device name has to be blank as well. The Labview VIs will not recognize nicknames.
To answer any connection questions, the device works just fine with Yokogawa's viewer software and programmer. It's only the Labview integrations that were having issues.
08-11-2022 06:45 AM - edited 08-11-2022 06:48 AM
some settings need to apply in gm10 web service ethernet configuration.. maintain your pc IP address there. then this will work ...
03-25-2024 07:06 AM
Can you please send the SUB.VI shown in image, because i have issue with the create same VI for the communication with Yokogawa GM10 & LABVIEW , so please can you share this VI
06-20-2024 04:23 PM
Were you able to find any VI? I'm also working with yokogawa GM10