11-22-2017 11:34 AM
I'm attempting to control a Tektronix PWS4205 programmable DC power supply using the LabVIEW samples that came with the driver. Eventually I want to program the cycle On/Off time of the voltage output to a motor while logging current over time.
My current issue:
The VISA resource name for the instrument is displayed in the control. When I attempt to run any of the example VIs or just the Initialize.vi I am unable to communicate with the instrument.
Error displayed:
The ID Query failed. This may mean that you selected the wrong instrument or your instrument did not respond. You may also be using a model that is not officially supported by this driver. If you are sure that you selected the correct instrument and it is responding, try disabling the ID Query.
I am able to use SignalExpress to output voltage and measure current on the power supply, but not when using example VIs. I'm not sure why I can communicate using the same VISA resource name in SignalExpress, but not through the VI.
Other info:
LabVIEW 2009 SP1
Windows 7
Solved! Go to Solution.
11-28-2017 08:58 AM
Have you tried to turn off the ID query in the Initialize function in your instrument driver?
http://digital.ni.com/public.nsf/allkb/0C03C3CCEC9A45B8862569DF006CCD27
Also try using the NI driver : http://sine.ni.com/apps/utf8/niid_web_display.model_page?p_model_id=18346
11-28-2017 12:28 PM
This system is using the command set that was out prior to SCPI becoming the standard. As such, it will NOT respond to the SCPI standard of
*IDN?\n
but instead will respond to:
ID?\n