Vötsch VTL 6010 communicat​ion with RS 232

Solved!
Go to solution

Hello,
 
We are using Vötsch VTL 6010 climatic chamber and we want to control it via RS 232 in Lab View
The problem is the device is not responding for our commands we have done the following settings on the MINCONTROL LCD
 
Baud rate 19200/9600
Bus 01
tried ASCII-1 and ASCII 2 Protocols
configured the unit to External mode.
 
I am sending commands using VISA Control but there is no resaponse.IS there anything else I am missing .Kindly suggest me
 
Thanks in advance..

0 Kudos
Message 1 of 5
(4,508 Views)
You could have the wrong com settings, the wrong command, the wrong (or missing) termination character, the wrong type of cable (two major types), etc.

You have not provided the manual, the command you are sending, the type of cable, or the termination character you are using. Essential information.
0 Kudos
Message 2 of 5
(4,492 Views)

Thanx for the response

 

Regarding the Com settings I have set it according to the manual

 

"$01I<CR>"

This was the string which i have sent.

I have connected the serial db9 pin according to the pin diagram given.

 

Sorry for not giving the manual...

 

this time i have attached it below..:)

 

 

Download All
0 Kudos
Message 3 of 5
(4,477 Views)
Solution
Accepted by topic author Bharadwaz
If you sent <CR>, then you did not send a carriage return. You sent the text. In MAX, you have to use \r.
0 Kudos
Message 4 of 5
(4,459 Views)

Thank you... it is working with ASCII 1 protocol... that will do for me... 😄

 

Cheerss..!

0 Kudos