05-18-2020 05:37 AM
Hi. I Have Sartorius PMA Scale. It has usb port only. They work normally with the standard program from the manufacturer. I am trying to establish communication and get data from the scales using LabView. I tried to run examples from the library but I don’t see any data. I ask for your help in the form of a working example or advice.
05-18-2020 06:47 AM
05-18-2020 07:06 AM
05-19-2020 04:20 AM
I try to run the example "Get Weight" but an error 107380753 occurs. Settings described on screenshots.
What is wrong?
05-20-2020 08:53 AM
Hi
Finally had time to look at this. A framing error means that either the baudrate or other setting of the scale is different from your software.
In the software this is hidden in the initialize function, ope it up and see 1200 baud, etc
It is not very friendly to hide this in a subvi without a connector so you can change it from the main program.
Anyway you should check if your scale has the same settings or change the settings in the initialize vi.
05-21-2020 07:27 AM
Dear Albert, I tried changing the port settings, but the error still remains. How can I set the start bit.
Technical parameters of the port as described:
Interface onnection USB, type B (virtual serial interface)
Format 7-bit ASCII, 1 start bit, 1 or 2 stop bits
Parity Even, odd, no parity
Transmission rates 600 to 57,600 bps
Handshake Software or hardware
05-22-2020 07:12 AM
Hi! It is hard to be stupid. Bud-rate on balance was wrong. Thank you very much for your time. You was quiet right!