Hi,

 

I am trying to communicate with an old Sorensen DCS power supply (DCS100-12) via RS232. It has the M9 option and the manual is attached. I am using a RJ11 to DB9 cable as mentioned in the section 2.4. 

 

Before going to a Labview VI, I am first trying to communicate using the VISA Test Panel. I am using the following parameters as mentioned in the manual :

 

Baud rate : 19200

Data Bits : 8

Stop Bits : 1

Parity : None

 

If I start the power supply and query "*IDN?\r\n" I get both the answer and a timeout error (see image below). The next query will always timeout (even *IDN? ). 

 

GR_JJC_0-1654880471230.png

 

I have tried a lot of different setting with no success. In the other hand I tried to use a GPIB to USB adapter from another device and it works fine. 

 

Questions :

  • Any ideas why the RS232 communication always timeout ?
  • Why it give me a good answer only for the first query ?

 

 

 

0 Kudos
Message 1 of 13
(3,793 Views)

You seem to do what I would have done.

My only suggestion is to use only \r or \n instead of both.

 

greetings from the Netherlands
0 Kudos
Message 2 of 13
(3,786 Views)

Thanks for your help,

 

I have also tried the \r and \n alone without success.

 

I guess I will have to buy another GPIB to USB adapter. 

0 Kudos
Message 3 of 13
(3,732 Views)

This power supply is unknown to me, so I can't speak from experience.

 

However, after spending a few minutes to read the instruction manual, I wonder if the problem does not come from the use of the IDN? command (instead of ID?).

 

M9A.png

0 Kudos
Message 4 of 13
(3,730 Views)

I tried the ID? command without success as well (with the different termination characters mentioned above). It timeout as any other read command. 

 

As I understand from the manual, M9A and SCPI are just two different languages but the HW setup is the same for both of them (cf Section 2).

 

I could imagine that the issue comes from the wiring, but I still receive an answer via the first IDN? command just after the power supply startup. 

 

I wonder if this can come from the VISA Test Panel. Is there another simple tool that can perform the same kind of tests ?

0 Kudos
Message 5 of 13
(3,721 Views)

Try entering *STB?\n or *CLS\n prior to sending the *IDN?\n queries.  It may be that their is a status message awaiting.

Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
Message 6 of 13
(3,716 Views)

@GR_JJC wrote:

...

 

I wonder if this can come from the VISA Test Panel. Is there another simple tool that can perform the same kind of tests ?


LabVIEW Menu >> Help >> Find Examples... >>

 

Serial.png

Message 7 of 13
(3,695 Views)

I tried both command with \n and \r\n as termination character.

 

Similarly to the *IDN?\r\n command, the *STB?\r\n works only one time after the restart of the power supply and it send me both a timeout error and the answer "68\r". The *CLS\r\n timeout the first time (even after a startup) and the next *IDN?\r\n timeout as well.

 

GR_JJC_0-1655187745056.png

 

0 Kudos
Message 8 of 13
(3,691 Views)

The response to *STB? gives us some insight as to what the issue may be.

 

The response of 68 = 64 + 4 as the status byte is a binary representation which is described on page 3-2 of the manual. 

(64) ==> It seems that you are configured with the Service Request = ON.  This could be due to a couple of things.

1) Do you have DIP switches on the back of the supply?  If so, make sure that the first 3 (MSB) switches are set according to figure 2.3. (This device is responding as if the 2nd switch is in the ON position)

2) *SRE is set in software.  Try writing, "*SRE0\n" to the device.

(4) ==> Implies that there is an error message waiting in the buffer.  Enter the query, "SYST:ERR?\n" to find the response and compare versus the SCPI Error Codes which begins on page 3-4.

Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
Message 9 of 13
(3,681 Views)

Thanks a lot for your explanation on the binary representation. 

 

I don't get the link between 64 and the service request. On page 7-3 the bit n.7 in the SCPI Status Byte correspond to the "Operation Status Flag".

 

(64) Service request

1) The 3 DIP switches are set up as described on the Figure 2-1 (for DCS Power supplies). 

GR_JJC_0-1655222289204.png

2) I tried the "*SRE0\n" command but it doesn't change anything. I also tried to query "*SRE?\r\n" after a startup and got "0\r". So I guess that is not the issue.

 

(4) Error message

I have restarted again the power supply and query "SYST:ERR?\n". I got the following error: 

 

201,\sUnexpected\swarm\sboot\r

 

The manual mention :

201 - Unexpected warm boot
This error means that the M9 GPIB-side processor experienced a warm boot that was unexpected, and it may indicate an internal crash of the M9 processor.

 

I don't think this comes from a HW issue as I tried with three differents DCS power supplies and got the same error. Moreover the power supplies are working using the GPIB to USB module. 

 

 