I have a Tunics-Plus laser (with IEEE-488.1 GPIB Remote Control) which I connected directly to the PCI-GPIB board (using Labview 6.1,windows 2000 Pro). After starting the Measurement & Automation controller the laser is listed as a instrument with the correct GPIB address but the program gives the following message "Instrument did not respond to the *IDN? query". After that, I tried to communicate withe the NI-488.2 Communicator but I got the message "iberr=EABO" which indicates a time-out condition after read. During sending the *IDN? you cann see the laser is doing something but it does not respond. After that, I tried to set the GPIB propperties for this instrument (i.e termination method etc.) using the GPIB Configuration W
izard. I think the problem is to set the right termination method. The laser-manufactures writes "Each message string sent br the computer/terminal to the laser is ended with the line-feed character (LF, ASCII code10). However, this character can be omitted if the EOI control line is set to true". It would be very nice if somebody could give me a suggestion how to solve this probem i.e. which settings I have to choose. p.s. If I connect a Tektronics Oscilloscope to the GPIB board everything is working propperly.