03-28-2017 05:12 PM
I have a ThermoScientific Orion Star A211 pH meter which I am trying to communicate with using LabVIEW and the simple serial example VI. The manual says the command to send is GETMEAS. When I use Hyperterminal I have no problems sending the command and receiving the response. However in LabVIEW I only get an echo of the command and not the response string. Any ideas?
I did see this thread which seems to be exactly my problem but when I put in GETMEAS\r\n I still don't get any response. I am using LV2016.
Thanks
04-07-2017 10:03 AM
Are you able to send any other commands or is it just this one? Have you sent a *IDN? command? Do you have NI-VISA installed and if so which version? Could you also link the manual if you found it online? I can't seem to find it.
04-07-2017 10:20 AM
I actually just figured this out late yesterday after digging around some more. The issue was the response from the instrument was multiple lines and I needed to disable the termination character when configuring the serial port. I could see that the bytes at port was 119 but I was only getting back around 30 which was just the echo before the actual data.
03-15-2021 06:22 PM
Hello,
I have the same problem with this model of pH meter, Would you please tell me how did you solve this issue?
Thank you
03-16-2021 10:43 AM
Make sure enable termination character is F on initialization and wait about 1500 ms from a write command before reading back the data:
03-16-2021 02:38 PM
Hello Nasu,
thank you so much for you help, could you please upload the Visa file, then I can work on that. I am stuck on this and my project has stopped just because of that. I appreciate your help.