Hi.

 

I'm using a usb-gpib-hs to communicate with a Fluke 5720A through a self written code with pyVisa. I'm having trouble when i try to apply voltages above 22V, i'm able to put the voltage value in the calibrator, but i can't apply the operate command, the operate command takes the equipament off the standby mode.

0 Kudos
Message 1 of 4
(1,094 Views)

Hello,

Have You tried to configure IVI drivers ?  Or I'll suggest  to check the DLL driver for Your Fluke device.  

Regards.

0 Kudos
Message 2 of 4
(1,050 Views)

Well, you can configure a max voltage, at least for the 5700, so I assume same for the 5720. Maybe it's set in your gear?

 

I usually set this value while programming and working with DUTs that don't like 1kV 😉    that magic white smoke is usually hard to get back into them 😄

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 3 of 4
(1,047 Views)

Thank you for the replies.

 

I found out there's a configuration that inhibits the "OPER" command when some kind of flags are marked, the solution is simply clear the flag counter and go on.

 

 