02-01-2021 05:58 AM
Hi.
I'm using a usb-gpib-hs to communicate with a Fluke 5720A through a self written code with pyVisa. I'm having trouble when i try to apply voltages above 22V, i'm able to put the voltage value in the calibrator, but i can't apply the operate command, the operate command takes the equipament off the standby mode.
02-02-2021 09:39 AM
Hello,
Have You tried to configure IVI drivers ? Or I'll suggest to check the DLL driver for Your Fluke device.
Regards.
02-02-2021 10:53 AM - edited 02-02-2021 10:57 AM
Well, you can configure a max voltage, at least for the 5700, so I assume same for the 5720. Maybe it's set in your gear?
I usually set this value while programming and working with DUTs that don't like 1kV 😉 that magic white smoke is usually hard to get back into them 😄
02-03-2021 05:33 AM
Thank you for the replies.
I found out there's a configuration that inhibits the "OPER" command when some kind of flags are marked, the solution is simply clear the flag counter and go on.