Not able to establish serial communicate with the use of the library "age364xa Getting Started.vi "
The error is shown as attached.
Solved! Go to Solution.
12-05-2008 08:23 AM
Hello Gelb,
I recommend that you use the VISA Interactive Control in MAX (Measurement and Automation Explorer) to troubleshoot the communication problem. Here is a good link for your reference: http://zone.ni.com/devzone/cda/tut/p/id/4370
Also, please let us know which instrument driver you are using and in what version of LabVIEW.
Cheers,
NathanT
12-10-2008 02:23 AM
Thank you nathan for your reply.
It is useful with the use of MAX.
It was discover that it requires the use of null modem cable for serial communication.
12-10-2008 07:45 PM
Hello Gelb,
I am glad to hear that you found the problem.
Cheers,
NathanT