06-07-2018 04:31 PM - edited 06-07-2018 04:52 PM
Hello, I´ve been trying to control a power supply from my pc using rs 232.
manual here http://www.us.tdk-lambda.com/hp/pdfs/Product_manuals/83515000.pdf
With it I set up the correct serial settings; Baud rate,data bits, etc.
I download the drivers of the company (even if my model isnt there I was told they would work just fine), as you may know there´s a example there but even while appearing to work in code it doesnt irl. It doesn´t do anything. ni drivers
I´ve used NI MAX to try to communicate, to test first (but I already have a program to send commands). By doing so I´ve gotten this error;
after the command it says
Write Operation (\sADR\s06\r\s)
No Error
Read STB Operation
Error
VISA: (Hex 0xBFFF0015) Timeout expired before operation completed.
I believe this is due to a no response.
I tried using the method in this thread, the last comment: https://forums.ni.com/t5/Instrument-Control-GPIB-Serial/Serial-Communication-with-TDK-Lambda/td-p/29... with no success.
I have found in other similar threads that it may be due to the cable, I use a rj45-db9 and db9-usb, I have use succesfully a db9-db9 and a db9-usb, and not the original cable. I´ve used this same setup to communicate to a multimeter.
06-07-2018 04:51 PM
It should not be a VGA cable. Those are 15 pin cables. RS232 is typically a 9 pin cables.
06-07-2018 04:53 PM
excuse me,it was a db9 I edit it
06-07-2018 05:57 PM
Hello,
You should Be using the cable (P/N: GEN/232-9) listed on page 74. And selected the rear switch:
"Select the desired interface RS232 or RS485 using rear panel setup switch SW1-6 (Section 4.4). RS232: DOWNposition RS485: UPposition b) Connect rear panel IN connector to the controller RS232 or RS485 port using a suitable"
I think that you Can use a simple serial terminal to test the communication. You don't really need To be worried with that now.
06-08-2018 08:17 AM
A couple of notes:
ADR 06\r STT?\r
Example response: MV(45.201),PV(45), MC(4.3257), PC(10), SR(30), FR(00) |
Normally, I would say that each command for SCPI should end with the \n (Newline) character. Here might be different.
06-08-2018 09:29 AM
Hello, thank you for answering, the switch is in that position, and the cable is in the IN as well. Do you have some other idea of what it could be? I'm baffle.
I believe I'm using the correct cable, but it was given to me from someone of IT. I'll show some pics.
Thank you
06-08-2018 09:34 AM
Hello, thank you, I tried your commands, but it still gives me the same error, do you have any other insigth?
When I work with a multimeter I had no problem communicating.
06-08-2018 09:58 AM
Hello,
Use a miltimeter and the pinout on page 74 to be sure. It's a waste of time the other odeas if the cable it's not correct.
I believe that you must send a termination char. Are you doing that?
06-08-2018 11:42 AM
@Jorge_Amaral ha escrito:
Hello,
Use a miltimeter and the pinout on page 74 to be sure. It's a waste of time the other odeas if the cable it's not correct.
I believe that you must send a termination char. Are you doing that?
I believe the guy from IT gave me an incorrect configuration for the cable, I'm already correcting it, I'll let know what happens.
I was sending it at first, but stop doing as it say so in the thread I posted.
06-08-2018 11:51 AM
I concur with Jorge, it looks like the wiring of the cable is incorrect based on the picture you posted of the RJ-45 end.
-AK2DM