03-22-2006 09:32 AM
03-23-2006 01:26 PM
Hi milan,
Is this the same computer with the same drivers as before except for the new -HS? If so you might be using an older version of the NI-488.2 driver that doesn't support the -HS. Here's a link to the current version of NI-488.2 that adds support for the GPIB-USB-HS. Just for completeness, here's a link to the current version of NI-VISA as well.
If you have installed the correct driver, the issue may be that when you first install a new device to the system, you need to run Measurement & Automation Explorer (MAX) and scan for instruments to generate the VISA resource descriptors for your devices.