Thank you all for taking the time to answer this question.
I am trying to control an Omega CNi32 series temperature controller with Rs232 serial communication enabled using Labview 2009SP1 through a GPIB-RS232CT NI converter
The connection is as follows:
1. From the PC to GPIB-RS232CT NI converter using a USB to GPIB cable
2. From GPIB-RS232CT NI converter to temperature controller using a DB9 cable
I can successfully control the temperature controller directly from the PC using a serial connection through the DB9 cable, but I need to use the GPIB-RS232 converter module for my application.
GPIB-RS232CT NI converter configuration: G mode with the address set, I can detect the GPIB address but couldn't communicate with the serial device.
The temperature controller is programmed to continuously transmit data and I tried using Visa Read and GPIB Read to get the bytes from the device, but no success.
I have attached the Omega temperature controller and GPIB-RS232CT NI converter manual for reference
Solved! Go to Solution.
01-30-2024 11:24 AM
@Hariram96 wrote:
The connection is as follows:
1. From the PC to GPIB-RS232CT NI converter using a USB to GPIB cable
2. From GPIB-RS232CT NI converter to temperature controller using a DB9 cable
Why? Just use a USB-RS232 adapter. Removing parts often simplifies the issues.
01-30-2024 11:39 AM - edited 01-30-2024 11:45 AM
@crossrulz wrote:
Why? Just use a USB-RS232 adapter. Removing parts often simplifies the issues.
You are right. However, my application demands the use of the serial to GPIB converter.
Info about my specific use case:
I am building a setup with multiple instruments that communicate with a PC using GPIB
The design is to use a single USB port in the PC and connect all the Instruments through GPIB to GPIB cables and one GPIB to USB cable to the PC.
I do understand I can use Mutilport UBS extenders to solve my issue. but I cannot use it for my design. I have to convert the serial device into a GPIB
02-01-2024 10:54 AM
I got it to work.
02-01-2024 11:27 AM
@Hariram96 wrote:
I got it to work.
- I tried to communicate with the instrument(the 2nd address indicating the serial device) using Ni max initially but was getting a timeout error message on the read command.
- The reason for the time-out error was due to the byte length I was trying to read
- By default it reads 1028 bytes which were not available and hence the time-out error
- Once I changed the byte size to the desired length I was able to communicate with the instrument without any issue
- Finally, I used the Visa read-write function to create a Labview Vi
Depending on the temperature chamber, you likely just needed to enable the Termination Character and then VISA Read would stop reading when it is read.