NI max cannot find VISA resource name for NI GPIB USB HS device

Solved!
Go to solution

Hi guys

 

I want to use labview to connect Agilent PNA Network analyzer using GPIB cable.

 

Labview cannot found visa resource name of it, just shows like below. I tried ways of reinstalling NI 488.2 and restart the NI GPIB Hardware Enumeration service, but still not working.

 

Any thoughts?

 

Thanks

ni.PNG

0 Kudos
Message 1 of 5
(2,638 Views)

First the stupid things:

 

1. Is the PNA set up to communicate over GPIB?

2. Did you hit the "Scan for Instruments" button?


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 2 of 5
(2,630 Views)

Hi, thanks for the reply!

 

The GPIB setting on PNA has no problem.

When I hit on scanning for instruments, the system shows me cannot find any instruments. 

0 Kudos
Message 3 of 5
(2,615 Views)
Solution
Accepted by topic author zyb1003

I found the problem. There are 2 GPIB ports on the back of my VNA. One is for connecting external device, one is not. 

0 Kudos
Message 4 of 5
(2,543 Views)

@zyb1003 wrote:

I found the problem. There are 2 GPIB ports on the back of my VNA. One is for connecting external device, one is not. 


Ah, you have one of those.  Yes, I have had that problem myself.  One of the reasons I prefer using the Ethernet connection instead of GPIB.


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos