Hello,
I'm trying to connect my laptop and Klinger MC2 by using GPIB-usb cable.
In NI-VISA, the instrument only can work well for "write" commend, but can't work for "query" and "read" commends. Does anyone know how can I fix it?
Thank you in advance.
-Jung
08-31-2021 10:22 PM
Verify the termination character used for query commands and GPIB settings.
Any settings on the instrument that are required to allow it to be controlled over GPIB. Please share a snippet of the GPIB settings you're using and a copy of the instrument's programmer manual.
Does the instrument come with any example program to control it over GPIB? did it come with LabVIEW drivers? do those work?