From Friday, April 19th (11:00 PM CDT) through Saturday, April 20th (2:00 PM CDT), 2024, ni.com will undergo system upgrades that may result in temporary service interruption.

We appreciate your patience as we improve our online experience.

Instrument Control (GPIB, Serial, VISA, IVI)

cancel
Showing results for 
Search instead for 
Did you mean: 

NI-Scope Soft Front Panel trigger turn off

Hello, I'm using a Ni 5122 high speed digitizer with the Ni-Scope Soft front panel. Is there a way to completly turn off the trigger? If I use "Immediate Triggers", the soft front panel acquires data for certain TIME/Divs-configurations... but if I increase the seconds per div.... the programm seems to wait for an trigger pulse? and no evolution of a for example wave is displayed Thanks
0 Kudos
Message 1 of 2
(2,684 Views)

Hi, we´ve tested the Scope SoftFrontpanel and found that, in immediate mode, only if the ref-position is being altered, the data acquisition will stop,

which is ok, because you´d like to move within a acquired buffer of values and don´t need new ones. When the ref-position is set back to zero, the scope will trigger again and take samples.

 

best regards

 

Marco Brauner NIG

0 Kudos
Message 2 of 2
(2,674 Views)