02-23-2021 05:09 AM
Hello
We have E5810B LAN / GPIB / USB Keysight gateway When we use their soft keysight connection expert 2020.
We can see all the equipment plugged in and used.
But when we do not connect to NI-MAX, we only detect the GPIB port (TCPIP0 :: xxxxx :: gpib0,8 :: INSTR). the usb and com ports are invisible.
Is there a specific driver to install?
02-24-2021 02:40 AM
Did you enabled NI-VISA Passport for Tulip or set NI Visa as Primary?
08-17-2021 10:54 AM
Hei, thank you for your answer. I have the same problem. Since I followed your instructions, the USB1 device is visible, but I can't open a VISA connection. In Keysight IO the device works fine.
Do you have any ideas?
04-05-2022 04:27 AM
Same issue with and without NIVisaTulip enable
Ni Support ?
04-05-2022 08:19 AM
Hi
After testing some solutions, a workaround seem work fine, when we
- Enable NI-VISA Passport for Tulip
- And change in C:\Windows\SysWOW64 NI Visa32.dll by keysight Visa32.dll