05-28-2020 10:01 PM
I have a GPIB device connected to my system, but I get an error when I try to scan for instruments using *IDN? command. The description says: "Binary Information received" After instrument timeout error is dispayed please find the attached image FYR.
Suggest is NI-GPIB card is defective or it can be solve by doing s/w update.
06-01-2020 02:56 PM
It might help to tell us what GPIB device you are using. Is the instrument SCPI compatible (should it respond to *IDN?)? Did you use the right termination character(s)?
It seems like your NI-GPIB device is working, and that you got a response. What isn't clear is what you were expecting or what the instrument should return.
More details would help.
Craig
06-02-2020 12:42 AM
Hi Craig,
GPIB controller is connected to R&S FSQ for the same device if i connected another GPIB controller am getting response from device please help me on this.
Kiran