Hi jmni:
Yes the VI_ERROR_BERR occurs when a GPIB EBUS error occurs. This will happen when there are no instruments at all connected to the GPIB board. A different error (no listeners error) occurs when there is just some incorrect addressing. That being said, the first thing I would do would be to check the cabling. Are your cable lengths within spec? Do you have other cables around you can switch out with? If the cables are all good, it would lead me to believe there is some issue with the instrument. The specified timeout should not have an effect on this error as again, it will only occur when there is nothing connected to the GPIB board.
Regards,
Emilie S.
10-07-2005 12:39 PM
Thanks for the feedback, Emilie.
I've seen the problem with two different GPIB buses with 2 different Boonton power meters. In both cases the cables have been 2 meters or less. I've also been in touch with Boonton themselves but they have been unable to shed any light on it.
jmni
10-10-2005 05:27 PM
Hi jmni:
Is Boonton able to reproduce the problem? Do you have any other instruments you can try with this GPIB interface? Any other GPIB interfaces you can try?
Emilie
10-11-2005 09:54 AM
Emilie,
The original configuration had a number of different GPIB instruments ( DMM, Oscilloscope, Waveform Generator, etc) on the same bus as the Boonton and none of them experienced the same problems. The Boonton was isolated onto its own GPIB bus so that it didn't interfere with the others and we still saw the problem. I have a test setup dedicated to another.Boonton and still see the problem. It's intermittent and it can happen on any command type. I've seen it happen within the first 20 commands sent to the Boonton and I've also sent as many as 18,000 commands to it before it occurred. There doesn't appear to be any way of regaining communication with the instrument without an Interface Clear.
Boonton have tried to be helpful but they really haven't seen this problem before. I've had some success with my work-around but it is by no means ideal.
jmni
10-12-2005 04:08 PM
Hi jmni:
Unfortunately I do not have a Boonton device here to test. It seems that since you can communicate to other instruments that the GPIB interface and the cable are both functioning properly. This leaves only the instrument to be not working properly. Has Boonton been able to test with the same NI GPIB interface that you have?
Emilie