That would be a nice feature. I've also got a misbehaving instrument. When it gets an *IDN? query, it locks up even though it is a supported command. This is not a fault in MAX, it is a fault in the instrument and it sounds like the HP instrument has some problem. No instrument should lock up if it gets an invalid query.
 
With this misbehaving instrument, I wish I could avoid running a scan for instruments but it's really not possible in my case. When an instrument is returned from cal, the first thing the technician does is a scan for instrument to verify that the instrument is set up correctly. We had to create a special procedure for this system to have him reset the goofy instrument after doing the scan. Having a setting in MAX to skip certain addresses would have made things a lot easier.
0 Kudos
Message 11 of 12
(1,025 Views)

This is a rarity, but I have heard that a bad GPIB cable can cause multiple instruments to be found in MAX like you are seeing.  If you have another cable handy, try that out just in case.

Chris R.

Chris R.
Applications Engineer
National Instruments