The device will not be recognized by max because it does not respond to *IDN, in fact this is an scpi command and the instruments are older than scpi.
BUT... MAX should not have more than one line that is not recognized, I normally get just one line with an instrument that old.
Can someone from NI respond ...
03-06-2006 09:16 AM
It appears that your instrument at PAD 5 doesn't respond to *IDN?, which is okay. The reason it shows up so many times is probably because it supports secondary addressing (think of it as a sort of subaddress). So you have an entry for PAD 5, SAD 96, one for PAD 5, SAD 97, one for PAD 5, SAD 98, and so on. You can check your documentation to see that your instrument does use secondary addressing.
Scott B.
GPIB Software
National Instruments
03-06-2006 01:46 PM
The 5386A is one of a series of frequency counters. The others are the 5384A and 5385A.
The manual for the 5386A is available from the Agilent site at:
http://cp.literature.agilent.com/litweb/pdf/05386-90010.pdf
The manual for the 5384A/5385A is available from the Agilent site at:
http://cp.literature.agilent.com/litweb/pdf/05384-90006.pdf
There is no mention of any sub address in these manuals.
Using the National Instruments "Interactive Control" utility the 5386A recognizes HPIB commands and responds.
The 5386A does not seem to interferer with other instruments on the bus.
Why does MAX look for subaddresses on this device?
Is it possible to prevent MAX from looking for subaddresses?
03-07-2006 04:18 PM - edited 03-07-2006 04:18 PM
MAX looks for subaddresses because each of the secondary addresses could represent a distinct "instrument" in one large box, all placed at one primary address. It is possible that you would want to send a command to a specific sub address. It's not possible to disable this feature, but I don't know why you would need to. We intend MAX to be a "get up and running" sort of utility for GPIB to make sure your instruments are detected. Yours clearly are. I do see that it does a poor job of indicating that the devices are recognized at secondary addresses. I'll see what I can do to change that in the future, but as far as I understand you're all set for now, unless you have further questions. Let me know.
Scott B.
GPIB Software
National Instruments
Message Edited by ScottieB on 03-07-2006 04:23 PM
03-07-2006 04:30 PM
No as a matter of fact I am not "set for now" because my instrument still cannot be recognized my MAX.
It seems that the VI for the frequency counter does not seems to work with when MAX declares a primary address as non-functional.
Is there any way for MAX to recognize that a primary address is in use by a non 488.2 instrument?