(Hex 0xBFFF00A5) The interface type is valid, but the specified interface number is not configured.

Hi everyone.
I'm a novice in using Labview. So here is the problem.
I have a device that communicate with computer through gpib to usb.
I've installed all drivers also Aglient IO control. Using Aglient I can communicate with device.
But when i tried to use in Labview I have such error.
When I change all VISA into 488 I have an error
Error 7 occurred at GPIB Write in test.vi.
I know that not so much information but I would really apriciate for help.
Thank you
Something is rotten in the state of Denmark
0 Kudos
Message 1 of 2
(4,986 Views)
Hello,
Can you tell me whether you installed Agilent VISA and NI VISA?

Also can you use the VISA interactive controls to query your device?
If not try enabling the Tulip.dll in the VISA passort.
Christian A
National Instruments
Applications Engineer