HP 4155A High voltage detected error

Solved!
Go to solution

Hello.

I have a question about the CAUTION in HP 4155A instrument.

 

We have a problem to control the instrument because of the error in the picture below.

 

Before the error occurs we used the instrument well by GPIB0 communication.

 
0 Kudos
Message 1 of 5
(1,405 Views)

No error attached. 

 

Can you perform the same task manually? 

 

I suspect you have not setup the interlock properly.  See - https://forums.ni.com/t5/Instrument-Control-GPIB-Serial/HP4155A-Fixture-Open/m-p/4056371#M81533

 

Craig

0 Kudos
Message 2 of 5
(1,392 Views)

Oops, thank you for your comment.

I attach the picture of the error.

 

0 Kudos
Message 3 of 5
(1,385 Views)
Solution
Accepted by JJ.Lee

Yikes, that's not an operation error but an SMU error from the calibration/diagnosis page.  I expect it appears even if you cycle the power?

 

Looks like the SMU has been damaged and needs repair.  You should contact Keysight for support.

 

Good luck.

Craig

Message 4 of 5
(1,344 Views)

Yes, it appeared right after when I turn on the power.

 

Thank you for your help.

I'll contact  Keysight for support.

 

Thank you so much.

Jea Jung Lee.