DMM Over range Warning but IsOverRange false

Solved!
Go to solution

Hi,

the function niDMM_Fetch sporadically returns the warning code 1073356801 (0x3FFA2001 "Over range.") with a measurement of 105 to 140 MOhm when using Auto Range. However the subsequent check with niDMM_IsOverRange returns VI_FALSE. As I understood 

 

Can anyone explain this or is this a bug?

 

 

Thanks in advance and best regards,

Michael

0 Kudos
Message 1 of 2
(870 Views)
Solution
Accepted by topic author NotSureJoe

So, for all those who are looking for the answer, NI R&D has confirmed this behavior as a bug (no. 2035720).

It occurs every time a resistance above 100 MOhm is measured with the PXI-4082 and Autorange enabled.

 

Best regards,

Michael