I need to be able to tare(zero) a scale with labview. The scale I'm using is the Mettler Toledo ML3002T. There are no drivers for this scale in labview. I also need to get the current weight displayed by the scale. Mettler Toledo scales use MT-SICS to send commands to the scales. Is it possible to be able to send those commands with labview to the scale?
10-24-2019 10:16 AM
I am guessing that MT-SICS is proprietary software provided by Mettler since you do not say what it is.
You can try to figure out what commands are being sent and the protocol is.
But unless they help you, you are just guessing.
10-24-2019 10:35 AM
I managed to find the command set for the Mettler Toledo Standard Interface Command Set (MT-SICS) here: https://www.mt.com/mt_ext_files/Editorial/Generic/7/MT-SICS_for_Excellence_Balances_BA_Editorial-Gen.... You should be able to do everything you need from there using a simple RS-232 interface.
10-24-2019 02:16 PM
I'm using a USB connection to the scale. If I use a VISA read to get a buffer reading it will display the stable weight, but I don't know how to write the commands from the MT-SICS in labview to the scale. I tried to use a VISA write to send the MT-SICS commands.
10-24-2019 02:26 PM
@ErickCapistran wrote:
I tried to use a VISA write to send the MT-SICS commands.
That is the function you should be using. But do make sure your commands are exactly correct, including the Carriage Return and Line Feed to terminate the commands. If your string is in "\ Codes" display, it will end with "\r\n". Or you can use Concatenate String with an End Of Line Constant (assuming Windows).