Control Mettler Toledo MS6002TS/00 scale with Labview using USB cable

Solved!
Go to solution

I'm able to send a simple command like Tare or send weight with the VISA test panel, but when I try to write a VISA command in Labview nothing happens. I'm ending the line with \r\n. I'm not getting an error using the VISA write to send the command. I have used this method before using labview to send commands to a Mettler Toledo scale but, I was using a RS 232 cable and I'm now using a USB cable. 

0 Kudos
Message 1 of 5
(1,317 Views)
Solution
Accepted by topic author ErickCapistran

Did you select \mode for strng display?

You can do this by right click the string and select that mode.

greetings from the Netherlands
0 Kudos
Message 2 of 5
(1,290 Views)

I did not select the mode. After selecting the mode code I was able to commutate to the scale. Thank you for answering my question. 

0 Kudos
Message 3 of 5
(1,282 Views)

Hello,

 

Could you share the VI or the list of commands? I have Toledo with LAN as interface and I want to trial it

0 Kudos
Message 4 of 5
(724 Views)

Mettler Toledo Scales have a set of commands "MT-SICS". You can google MT-SICS to get the list. The commands you can use are dependent on the model of your scale, the MT-SICS manual has the list. I used VISA commands for my code, at the end of the command write "/r/n", and make sure the string is set to code.

0 Kudos
Message 5 of 5
(705 Views)