Bug Report: Tek DPO 4000 oscilloscope driver (won't Recall Setup)

There's an error in this instrument driver: tkdpo4k\tkdpo4k.llb\tkdpo4k Save-Recall Setup.vi

 

Save Setup works but Recall Setup doesn't.  On the block diagram, Recall case, the 'Initial String" input constant to "Format Into String" vi is "*RCL"  It's missing a trailing space as in the Save case "*SAV ".

0 Kudos
Message 1 of 4
(2,939 Views)

Me again...same Vi.

While we're at it, the addressing of the location to save and recall setup is off by one. The old 0 vs. 1 index for the first value of an array.  The setup storage locations for this scope are 1 through 10.  This 'tested' instrument driver tries to address them as 0 through 9.

0 Kudos
Message 2 of 4
(2,938 Views)

Hi chris8332,

 

Thank you for the feedback. We will try to reproduce this issue on this side and check it out.

 

Thank you!

Chris T.
0 Kudos
Message 3 of 4
(2,903 Views)

Hi chris8332,

 

This was reported to R&D (# 285920) for further investigation. Thank you for the feedback!

 