Hi all,
I want to control my Keithley 6485 Picoammeter externally by connecting it via RS232toUSB to my Linux PC (CentOS 6.9) and writing a code in python (version 2.7.13) with pyvisa:
import sys
import visa
from visa import constants
rm = visa.ResourceManager('/usr/local/vxipnp/linux/lib64/libvisa.so')
#open serial connection and set baud to 9600, 8 data bits, CR termination, one stop bit, none parity, no flow control
amm = rm.open_resource('ASRL2::INSTR', baud_rate = 9600, data_bits = 8, write_termination= '\r', read_termination = '\r')
constants.VI_ASRL_STOP_ONE
constants.VI_ASRL_PAR_NONE
constants.VI_ASRL_FLOW_NONE
amm.write("*RST") # Return 6485 to RST default
amm.write("SYS:ERR:ALL?") # Return error message
amm.write("TRIG:DEL 0") # Set trigger delay to zero seconds
amm.write("TRIG:COUNT 2500") # Set trigger count to 2500
amm.write("SENS:CURR:RANG:AUTO OFF") # Turn auto range off
amm.write("SENS:CURR:NPLC .01") # Set integration rate to NPLC 0.01
amm.write("SENS:CURR:RANG 2e-7") # Use 200 nA range
amm.write("SYST:ZCH OFF") # Turn zero check off
amm.write("SYST:AZER:STAT OFF") # Turn auto zero off
amm.write("DISP:ENAB OFF") # Turn Display off
amm.write("*CLS") # Clear status model
amm.write("TRAC:POIN 2500") # Set buffer size to 2500
amm.write("TRAC:CLE") # Clear buffer
amm.write("TRAC:FEED:CONT NEXT") # Set storage control to start on next reading
amm.write("STAT:MEAS:ENAB 512") # Enable buffer full measurement event
amm.write("*SRE 1") # Enable SRQ on buffer full measurement event
amm.write("*OPC?") # operation complete query (synchronize completion of commands)
amm.write("INIT") # start taking and storing readings wait for GPIB SRQ line to go true
amm.write("DISP:ENAB ON") # Turn display on
print(amm.query_ascii_values("TRAC:DATA?")) # Request data from buffer
The problem when I run this script I just get "1" as the print output, although it should be returned in ASCII like this: Reading, Timestamp, Status and the error message after amm.write("*RST"): -113 undefined header. So I think the messages and not transferred correctly.
I know over the RS-232 interface, only the ASCII format is allowed. But when I follow the example in the pyvisa instruction with write_ascii_values(text, values) and assigning it a list, I only get an error message from the device -100 Command error.
Can somebody please tell me how to set the variables in write_ascii_values correctly or what I am doing wrong? Are my settings for the serial device wrong? Sometimes when I execute 2 times I get the error "VI_ERROR_ASRL_FRAMING (-1073807253): A framing error occurred during transfer ." too. I just do not know what to do.
Thank you!
Regards, Roland
08-02-2017 06:42 AM - edited 08-02-2017 06:52 AM
Try adding a "\n" to every command and query.
amm.write("*RST\n")
Also, how are you handling the queries? All I see are amm.write commands. Where are the amm.read commands following a write command with a "?" within the quotes? Once you send a command containing a "?" in the message, the instrument will then want to respond to that request. The "1" that you receive in your print function is probably from the earlier command
amm.write("SYST:ERR:ALL?)
I do know that when you send the
amm.write("*OPC?")
command that it will respond with a "1" to signify that the instrument has completed processing all earlier commands.
08-02-2017 08:06 AM
Hi,
thanks for your answer. The printed "1" comes indeed from the command
amm.write("*OPC?").
When I erase OPC? and SYS:ERR:ALL? from the code I get the error:
File "./AMDAQ.py", line 31, in <module> print(amm.query("TRAC:DATA?"))
pyvisa.errors.VisaIOError: VI_ERROR_TMO (-1073807339): Timeout expired before operation completed.
The adding of "\n" did not help (still get the -113 error). I assign the read and write termination as "\r" here:
amm = rm.open_resource('ASRL2::INSTR', baud_rate = 9600, data_bits = 8, write_termination= '\r', read_termination = '\r'
Adding \n would lead to \n\r, so a LFCR termination for the device. However I tried every termination possibility with the device: CR, LF, CRLF, LFCR, but it does not change anything.
08-02-2017 10:18 AM
I am not that familiar with this instrument, but typically, the first thing to do after a reset is to configure the instrument for the type of measuring.
CONF:CURR:DC
Also, did you try using "READ?" for your print function
print(amm.query("READ?"))
08-02-2017 10:35 AM
The manual is confusing as it sometimes refers to the same command with or without the prefix of ":"
:SYST:CLE or SYST:CLE
Not sure if it matters?
08-02-2017 01:38 PM
BTW: Keep the *OPC? statement where it is, just change it to a query as it allows the system time to configure itself properly before responding.
amm.query("*OPC?")
08-03-2017 08:06 AM
Thanks again for your help. I amended my script according to your change advises. It does not change anything.
I also change from RS232 to GPIB, but I get the same VI_ERROR_TMO (-1073807339) error. Also when I change the language from SCPI to 488.1. So there is definitely something wrong with the translation when it wants to read the data out of it. But I am out of ideas...