0V & 0A DC/AC Negative Tolerance (34410A / 34401A / 34465A)

Note: Duplicate of https://community.keysight.com/message/91866

 

What are the negative fluctuations that we expect with preferably the 34410A (or the 34401A) when we are in an open circuit or when a device is turned off for DC&AC Voltage&Current measurements?

 

I want to avoid the -0 issues with automated measurements, however I need to have a valid reason to put a negative value as a low limit.

 

Measurement Range Accuracy Math Tolerance
DC Voltage 100mV 0.0050+0.0035 0.000035*0.1mV 3.5 uV

AC Voltage

100mVrms 0.5+0.03 0.0003*0.1mV 30 uVrms
DC Current 100uA 0.05+0.025 0.00025*100uA 0.025 uA??
AC Current 100uArms 0.1+0.04 0.04*100uA 0.04 uA??

 

If I consider full range for reading:

Measurement Range Accuracy Math Tolerance
DC Voltage 100mV 0.0050+0.0035 0.000085*0.1mV 8.5 uV

AC Voltage

100mVrms 0.5+0.03 0.0053*0.1mV 0.530 mVrms
DC Current 100uA 0.05+0.025 0.00075*100uA 0.075 uA
AC Current 100uArms 0.1+0.04 0.14*100uA 0.14 uA

 

Can anyone confirm or correct? Especially for AC since the specs are for at least 0.3% of range and a frequency over 3Hz.

 

Thanks

0 Kudos
Message 1 of 6
(2,878 Views)

As this is a Keysight device, it's not something we can answer spec questions about and guarantee we'll be correct. Please rely on the Keysight forum for an answer to this question.

Lindsey Nestor
Program Manager, Hardware Services
National Instruments
0 Kudos
Message 2 of 6
(2,854 Views)

Your question isn't very clear.  What do you mean by "negative fluctuations", "-0 issues" and negative value as low limit? 

 

The DMM can measure both positive and negative voltage and current so the specs posted are most likely absolute values.   

 

Elaborate on what you want to know, or describe what you are seeing and perhaps I can offer a few comments as I use a 34401a quite often.  (Mainly in DCV and DCI modes.)

 

Craig

0 Kudos
Message 3 of 6
(2,845 Views)

Hi Craig,

 

Sometimes when no voltage or current is measured with the DMM it will return a very small negative value such as -0.000001A.

Try powering ON your DMM and ready the voltage with without connecting the leads (same as reading an open circuit) and you should be getting some negative values (even if you ground the low side probe).

0 Kudos
Message 4 of 6
(2,820 Views)

That becomes an accuracy issue of the instrument.  If you have ever dealt with an ADC then you would know all about INL & DNL.  The instrument should provide sufficient information in order to make a determination on how LSB's of accuracy then based upon the range chosen you can select your negative low side value. 

Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
0 Kudos
Message 5 of 6
(2,806 Views)

That gives some extra information to dig through specs. I remember using the INL/DNL for NI DAQ cards for tolerance calculations in the past. Although Agilent didn't provide that information 😛

 

From the specs I found extra information that I think I can add:

A-to-D Converter Noise Performance: @30ppm of range

DC Voltage Linearity: 0.0002% Reading + 0.0001% Range