Hello,
I use the softpanel for acquiring some data with a high-speed digitizer 5122.
Now some questions arouse.
1) The Time/Div value.... I can choose for example 5 seconds per DIV. ... But as my wanted signal is only about 36 second, I do on the one hand loose some time...
On the other side.... the acquisitian time is to short...
=> so is there any possibility to manuelly adjust the time/div ... (I also want to keep the 900.000 acquisition points)
2) Is there any possibility to save the measured signal in two formats .... lvm and hws? When I choose "lvm", and later hws... I don't come in the desired menu? (Or is this a case of "problem exists between keyboard an chair")
Thanks
Andreas
02-10-2012 10:45 AM
Andreas,
1) The available options on the Time/Div drop-down correspond to the available sampling rates for the specific device (See KB). The only way to change these options would be changing the record length (which is sounds like you don't want to do). However, if you just want the UI to focus on the specific section, there is a zoom mode available. If you click the magnifying glass in the upper right hand corner of the graph, it allows you to see the full acquired waveform on top and then use cursors to select the region you want to view below it.
2) I'm not sure if I fully understand where you are running into difficulty. If you choose to Save Data, after hitting okay it brings you to a menu where you can select the format. I tried saving in both HWS and LVM and they seemed to work (I opened them both in a text editor and the formats were what I expected). Where exactly are you running into trouble?
I hope this helps,
-Lauren