High-Speed Digitizers
PXI 5124 and PXI 5153 noise
Hello everyone,
I am using PXI 5124 (200 MS/s, BW:150 MHz) for acquiring photoacoustic (PA) signals ( PA signals are the same as ultrasound signals).
A transducer 75 MHz is used to detect the PA signals, then these signals are amplified (54dB) to PXI 5124. It's working well.
Now, I want to use PXI 5153 (2GS/s, BW:500 MHz), and noise appears. Please see more details in the attachment files
I use the same configurations for both PXI 5124 and PXI 5153 which are passed self-calibration.
Can you let me know what is wrong with my settings? is there any problem with my PXI 5153?
Thanks in advance
Regards
WAM
(3,044 Views)
Accepted by topic author WorldALoneMe
First 'obvious' deviation:
The lower noise 5124 has 12 bit resolution ... the 5153 has 8 bit. ..
next: even if you requested the same range, both cards have different ranges, so read the actual range property to check 😉  I guess you compare 12bit at 400mVpp and 8bit at 500mVpp  (one bit ~2mV !)
but  maybe the 5134 always sample at 2GS/s and mean the values while downsampling ... so RTFM 🙂
However the spec tells already
Signal to noise ratio : 5124: 58 db  to  5134: 45 db     , a 13db difference !!
Greetings from Germany
Henrik
LV since v3.1
(2,968 Views)
Thank you for your reply, Henrik.
I totally agree with you about the big difference in SNR.😂
In my application, the photoacoustic signal is very small. (It gained 54 dB before going into digitizer).
So I can not use PXI 5153 for my project (the noise is over the signal).
I analyzed the noise by using FFT, and I see that the noise frequency at 5 MHz. (please see more details in the attachment)
Do you know why 5 MHz?
Thanks & Regards
WAM
(2,953 Views)
5MHz noise: Also present if the input is only a 50 Ohm termination?
Since you use a preamplifier anyway , you can always add some db to get over that noise 😉
Henrik
LV since v3.1
(2,931 Views)
Currently, I config at 50 Ohm termination for sensor coupling (transducer, cable: 50 Ohm).
How can I add some dB in the preamplifier, Henrik?
I have no idea about getting over the noise by adding something (dB).
Can you refer me to some documents about this technique, please?
WAM
(2,923 Views)
If you measure the noise flor of the 5153 just with another 50 Ohm termination at the input (or without termination but a shielded open input and only the internal 50 Ohm) do you still see the 5 MHz?
You can try to add another amplifier between the preamplifier and the 5153 ...  add faktor 10 (20db)
(or turn the gain knob at you preamplifier 😉 )
Or if you have a periodic signal you can use lock-in techniques ...
Henrik
LV since v3.1
(2,913 Views)
There was no 5 MHz noise when I connect PXI 5153 with the 50 Ohm terminator.
The background noise 5MHz just appears when I turn on the preamplifier. (No noise with 5124).
So what could be the problem?
WAM
(2,909 Views)
