I have a test stand setup where two different DAQ boxes are being used, one with a cDAQ-9171 (and a NI-9862 CAN module) and one with a cDAQ-9178 (with multiple modules). The 9178 is to be interfaced using FlexLogger, and the 9171 has its own custom Labview executable to communicate with the CAN. I would really like to use a single computer to communicate with both of these, but when I plug them in, FlexLogger recognizes and reserves both, preventing the Labview program from reserving the 9171. If I start the Labview program first, it successfully reserves the 9171, but then FlexLogger "steals" the chassis away when it starts. Is there a way to disable the 9171 within FlexLogger, or some way to have Labview refuse to have it stolen away?
As a side question, I have a second setup in work that will use a USB-485/4 module instead of the cDAQ-9171. Will FlexLogger try to connect to the USB-485/4? I didn't see it in the list of supported devices for FlexLogger. I'm wondering if I'll have the same problem with that setup or not.
Solved! Go to Solution.
07-28-2026 04:14 PM
I ended up opening a technical support ticket for the above issue, and I was directed to the "Disable" checkbox in FlexLogger (see image attached). This allowed that chassis to be free to be reserved by my separate Labview program. I'm not sure how I missed that checkbox previously, but it fixed my problem. Kudos to NI tech support for their help.