FlexLogger
4 Wire Load Cell on NI9219
Evening All,
I’m looking for the best way to connect up a 4 wire load cell to our Compact Daq with a NI9219 universal module, using Flexlogger.
load cell info:
10v reference voltage
2mv/v sensitivity
500kN nominal load
Any ideas?
(2,006 Views)
4-wire load cell is a full bridge config, and here is how you must connect it,
Another variant of the connection diagram the DAQmx task,
Santhosh
Soliton Technologies
(1,987 Views)
Thanks, that is how I had it connected up, but the card will not allow for more than 2.5v ext voltage, unless I'm missing something.
Do you know which settings to use on the module setting screen?
(1,949 Views)
Yes, that is expected by the 9219 hardware design and described in the specifications,
Santhosh
(1,943 Views)
What would be the best module to use for this application? Based on the load cells parameters mentioned previous?
(1,937 Views)
@Dan_Buck wrote:
Evening All,
load cell info:
2mv/v sensitivity
500kN nominal load
Any ideas?
Does the load cell need such high excitation voltage? I would also recommend checking other load cells with a lower excitation voltage if those are comparatively cheaper than a new c-series module.
Santhosh
(1,934 Views)
9219 and 9237 are the only c-series modules that support full-bridge load cells.
9237 supports a variety of internal excitation voltages, including 10V.
Santhosh
(1,932 Views)