Program labveiw to read torque

Hi Steve,  
          Thank you for the info. you have provided so far.I have tried to reproduce the settings you have posted with limited results.I am using the AI-110 to measure voltage with a the range set to -0.065 to 0.065volts but when i try to excite the torque senor with 10vdc using ethir the FP-A0-210 or FP -SG-140 i get the following errors:  FP-AO-210= (channel is soucing the maximum current allowed) FP-SG-140=  (CHANNEL IS IN
0 Kudos
Message 11 of 12
(3,320 Views)
Hi Bruce,

The AO-210 will source 1mA without and 10mA with an external power supply (page 1).  The SG-140 requires a minimum bridge resistance of 477Ω at 10V of excitation (page 7).  You should check with Futek for the maximum current the device will draw at 10V of excitation.  If it's greater than 10mA for the AO-210, or 21mA for the SG-140, you cannot excite the sensor directly from FieldPoint module.  This is not a big problem.  In fact, it points to a better solution altogether.

If you excite the sensor with an external power supply (link a standard 120VAC - 10VDC converter), you could measure both the sensor output and the excitation voltage with the AI-110.  The computed torque would then be a function of two real-world measurements, not a real-world measurement and a software setting.  I'll assume that, "Now, the numbers i am getting from the AO-110, what do they represent?" is a typo, and that you're asking what the numbers from the AI-110 are.  As long as you have not applied any scaling to the AI-110 channel in Measurement and Automation Explorer (MAX), these numbers are voltage readings in units of volts.  Please see the previous post on how to compute the scaling from volts to torque in the LabVIEW program.


If you like, you can linearly scale the voltage data on a FieldPoint channel.  In MAX >> My System >> Data Neighborhood >> FieldPoint Items >> <Controller Name> >> AI-110 >> Channel n >> Scales.  You can specify a slope, a y-intercept, and scaled units.  The units in LabVIEW for such a channel would be whatever you set in MAX.



Message Edited by Pie56694 on 07-02-2008 02:19 PM