pxi-4110 dcpower soft panel cannot find device

I have a brand new PXIe-1085 system with two NI-PXI-4110 DC supplies.  The DCpower soft panel does not find the device.  I am using DCPOWER 15.1.  any suggestions?

0 Kudos
Message 1 of 15
(8,009 Views)
  • What operating system do you have (version, bitness)?
  • Do the boards show up in MAX?
  • Do the boards show up in Windows Device Manager?
Marcos Kirsch
Chief Software Engineer
NI Driver Software
0 Kudos
Message 2 of 15
(8,003 Views)

Windows 7 64 bit.  The devices are in MAX and the device manager

0 Kudos
Message 3 of 15
(8,001 Views)

Windows 7 64 bit.  The devices are in MAX and the device manager

0 Kudos
Message 4 of 15
(7,995 Views)

Could you check if ModInst and SysConfig recognize the device? 

You could use the two attached VIs (LabVIEW 2012).

Download All
0 Kudos
Message 5 of 15
(7,991 Views)

We use CVI

0 Kudos
Message 6 of 15
(7,984 Views)

Here is a CVI example for using NI-ModInst: http://www.ni.com/example/28726/en/

 

What happens when you call Self Test or Reset on either 4110 in MAX?

What happens when you call Soft Front Panel from either 4110 in MAX?

Are you able to create a simulated PXI-4110 from MAX?

Tobias
Principal Software Engineer
Driver Software
National Instruments
0 Kudos
Message 7 of 15
(7,973 Views)

the cvi code returns no devices found.

I cannot reset the device in MAX but it is present (do not see reset option (but can open a visa test panel)

MAX inidicates the device is ready

The dc soft panel starts but stops with an error message "no valid devices found"

windows device manager sees the devices but the driver version is not available.

 

0 Kudos
Message 8 of 15
(7,956 Views)

Hi Joncombat,

 

Were the NI 4110 modules both just purchased from NI?  

 

I've seen similar behavior from 4110 modules that were missing their user-replacable fuses.  Refer to this document for the location of the fuses and make sure that they are there.  You could also use a DMM to make sure that the fuses are good.  I would expect the resistance to read < 1 Ω.

Regards,

Jared R.
Precision DC Product Support Engineer
National Instruments
0 Kudos