05-29-2010 11:02 AM
06-01-2010 04:54 PM
Hello Magnus,
I believe you are asking whether the HI and LO outputs of the 4132 can be individually controlled as if they were independent power supplies. The way it works on the 4132 is that there are voltage and current control loops which sense the HI/LO differential voltage. When we program 100V, we measure back the output voltage (via the control loop) and adjust the output to match 100V. The same thing happens for -100V (i.e., the differential between HI and LO is equal to the programmed output).
For your application, I would recommend cascading two 4132 SMUs. By cascading the outputs of both SMUs you will essentially have one SMU ... but you can program the "HI" and "LO" independently of each other. I hope this is helpful, let me know if anything I stated in this post is unclear.
06-02-2010 02:47 AM
Thanks Rod,
I think that clarifies it.
I don't actually need fully individual control of each terminal, just to mirror each other in sign of potential. If the channels are isolated I guess I can make a simple voltage divider by a par of serial connected resistors and connect the middle section to ground and I will get a ±50V on each of the termianals, is this right?, se the sketch that I made. Of course I will only get a max potential of 100V in total with this configuration.
/Magnus
06-08-2010 10:57 AM
Hello Magnus,
Yes you are correct. Assuming you tie ground to the middle as shown in your diagram and measure from + to Gnd, you will see +50V. Measuring from - to Gnd you will see -50V.