Newb: DMM 4071 7.5 digit low sample rate

Apologies in advance for the newbie questions....

 

I am trying to se up a test where I need a PXI-4071 to run in it's full resolution 7.5 digit DMM mode at 7Hz?, ie not waveform mode.

 

Using (fully updated):

- .Net 4.0 / Visual studio Express 2012

- NIDMM .net class libraries 1.0.3

- Latest DMM driver that comes with soft panel ver 14

 

ContinuousAcquisition.2010.sln I get it running at 100hz or more when setting resolution to 7.5 digit

 

SoftwareTriggeredMultipointAcquisition.2010.sln, seems to run at around 1hz, also in imediate mode. Is this the "full 7.5 digit DMM mode"? Why not 7Hz?

 

I am so far running the code in Visual studio Express in release mode.

 

Can someone please point me in the right direction, ie documentation and sample code.

 

0 Kudos
Message 1 of 3
(5,770 Views)

 

Problem solved; there is an error in the demo code (ContinuousAcquisition.2010.sln)

 

Wrong:

sampleDmmSession.Configure(measurementMode, range, resolution)

Should be:

sampleDmmSession.ConfigureMeasurementDigits(measurementMode, range, resolution)

 Otherwise the module will run at high speed 3.5 digit resolution....

 