Hi

 

The NI DC Soft Front Panel V14.0, with the PXI-4110, the negative voltage scroll, works as expected to -10V, but then rolls over to 0. If one changes in -1V steps, it goes.... -8, -9, -10, -1, -2 ..... instead of -8, -9, -10, -11, -12 ...

 

While at it, a simple thing I miss is a on/off button for all three voltages.

 

(Also, imo, it would be logical to get negative voltages with the down arrow, not up).

 

 

My 2C

0 Kudos
Message 1 of 2
(5,770 Views)
Solution
Accepted by topic author janaf

Hello Janaf,

 

I fully agree with both of your statements, I have filed a Correction Action Report which you can monitor in next releases of DCPower to see if this is fixed with the SFP. Number of CAR: 512257

 

I added notes that only manual insertion of numbers less than -10 works and that it was not logical to use increment arrow or up arrow to decrease the voltage output.

Best Regards
Jonas Mäki
Systems Developer
Novator Solutions