08-02-2016 04:07 PM
The NI 408x DMM specification format has changed from the NI 407x series. This leaves some questions and lack of information on how to calculate the accuracies for ranges less than the maximum (e.g. 5.5 digits). There is a general lack of documentation for the new meters with few hits on the NI website.
Without complete specifications, I can’t determine the suitability of these meters for future use, especially if the NI 4071 is discontinued.
John Anderson
08-03-2016 09:06 PM - edited 08-03-2016 09:07 PM
Hey John,
I hope this information is helpful for you. I can see that there are still some gaps in the information we've reviewed so far, and I understand how that can make it difficult to determine if a specific device is right for your application. I briefly suggested this above, but considering the depth of your questions, I would suggest reaching out to our support team to open a ticket and review this information directly with an Applications Engineer. The AE you will work with can do some deeper research and consult our PSEs and developers as necessary. (You can reach the support team at 866-275-6964 or by emailing support@ni.com).
Take care,
Will
08-04-2016 12:11 AM
Hello Mr. Anderson,
Thank you for the detailed questions. Here is some clarification and rationale for the 408x specs.
In the past, we received feedback that the 407x specs were difficult to interpret and compare with other dmms. With the 408x, as you pointed out, we updated several aspects of the specs, considering both industry norms and aiming for increased consistency with our related SMU products.
You are correct -- for the 4081, DCV is now specified over the full 0-55C environmental range, though explicit specs are given assuming device temperature is within 5C of the last self cal. The footnote on page 5 (resistance accuracy over full temperature range) was included to differentiate that column from the other accuracy columns that constrained temperature to +/- 10C from External Calibration. Because the voltage reference is so stable, there wasn't much value in adding those additional columns to the DCV accuracy table.
For DCV, tempco is provided for guidance on device drift with or without self cal, but it is not necessary to add that back in if you stay within 5C of the last self cal.
We moved away from using absolute environmental temperatures (0-55C, 18-28C, etc.) for accuracy criteria, instead using device temperature (Tselfcal +/- 5C, Textcal +/- 10C, etc.). This is because DMM temperature can vary significantly depending on module load and activity in various PXI chassis. The onboard temperature is what matters - not that of the outside air.
In many cases, the 4071's primary absolute accuracy specs required +/- 1C, but we assumed most customers would have a hard time holding that, so we loosened the 4081's specs to allow for +/- 5 C. Relative accuracy specs, as is customary, remain at +/- 1C.
We documented aperture specs in absolute time rather than power line cycles because the it provides consistency among systems deployed in 50 Hz and 60 Hz regions… 100 ms aperture for a "6.5 digit" measurement has the same timing in the U.S. and Europe. The difference between 5 and 6 periods isn't very significant when windowed with the default High-Order Noise Rejection.
As a PXI instrument, we have the benefit of two isolation layers from AC MAINS to the analog front-end (an AC-DC conversion in the chassis and an isolated DC-DC in the DMM). Power line noise within the instrument is not dominant, though it obviously could be depending on system cabling.
The noise curves sweep aperture time from 555 ns because the 408x is now capable of acquiring that narrow a sample. The curve stops at 100 ms because you get diminishing returns beyond that. For very long apertures, you may see slightly increased drift because there hasn't been recent auto zero. Rather than using a long aperture to reduce noise, it is better to increase Number of Averages, which will repeat the Auto Zero stage.
For the noise multiplier, we used 5x not only to convert from RMS to peak, but also margin up from typical since you would be applying this scaling on top of a warranted accuracy. A noise parameter has already been included in the absolute accuracy specs, assuming a 100 ms aperture. You only need to account for additional noise if you use a shorter aperture.
For Digitizer mode on the 407x, DC accuracy was listed as "typical". For the 408x, we opted to provide warranted DC accuracy specifications. Because you can't interrupt a continuous waveform with correcting functions (like auto zero), it's not possible to achieve the same DC accuracy as the DCV DMM mode. DC Accuracy is not dependent on sample rate, but noise is. We provided the typical noise curves in RMS so the user could consider it in conjunction with other system noises (rss-ing them, for example). We did not attempt to convert to Peak (or Peak-to-Peak) because Digitizer mode is inherently a multi-point acquisition where DC accuracy applies to the average over many points. Compare to DMM mode which is often single-point, where the noise contribution needs to be considered as a peak to add to the accuracy.
Keep in mind that Digitizer mode's AC performance specs (flatness and bandwidth) are still "Typical", which is not unusual for this type of device.
The default aperture times and other configuration parameters associated with a given "number of digits" setting are defined in the driver. As such, they are documented in the NI-DMM Help. You can find it under:
NI Digital Multimeters Help -> Devices -> NI 4081 -> DMM Measurements -> DMM Measurement Defaults
I apologize for the long post, but wanted to give some context for spec changes between the 407x and 408x.
If you are at NIWeek, I hope you've found it to be interesting and educational. If you have further questions, please don't hesitate to ask.
Charles Yarbrough
08-04-2016 05:05 PM
Thank you for your responses and I have some follow up.
It sounds like digitizer accuracy depends on the actual measurement being taken. I will check with our experts to see how we will handle this.
The NI 408x devices do not show up under DMM help but all the other meters are there.
The meters look good but the spec sheets need some work. They don’t even specify how many digits can be used for each function. Example accuracy calculations are appreciated even if it is just a white paper.
John Anderson
08-05-2016 10:14 AM
The NI 408x was just released and is first supported in NI-DMM 15.2. Is this the version you have installed?
08-10-2016 11:45 AM
After a 6 hour ordeal to download and install NI-DMM 16.0 (upgrading from 3.1) I found the default documentation that looks the same as the NI 4071. But my complaint was the website help doesn't get updated to the latest information in a timely manor. In the past our IT security was blocking access to all help files so I had to go on line since pdf files have not been available for some time.
John Anderson
08-12-2016 04:26 PM
Hey John,
I'm sorry that you had a rough time with the newest driver, but hopefully now you are able to access either the online or offline help for NI-DMM 16.0, which should both be available to you.
When you mention our online information are you referring to the help documentation or to other articles such as KBs, white papers, etc.? If so I can tell you that our content teams are actively seeking to improve our process for keeping our resources up to date, so that when we release new products, our information can be updated as quickly as possible (it's a pain point for us too).
But certainly as always if there is ever a question you have, or clarification you need, you can either leave feedback on the resources themselves, or give our support line a call and our AEs can help you track down what you need.
So just to clarify have you now been able to find the clarifications you needed from our 408X documentation, or do you still have some outstanding questions?
Best Regards,
Chris J