04-26-2022 07:18 AM
To self-calibrate a board using NI-DMM, all you need to do is:
Be aware that your board's external calibration is expired, so running self cal will not get it to specification, and may or may not succeed at all. I suspect you will get the same error you got in MAX. I recommend you find a service to perform external calibration on your instrument.
04-26-2022 07:51 AM
04-26-2022 07:58 AM
No. To be within specs the DMM needs to be externally calibrated before the calibration expires. Self calibration is recommended to be run more frequently, I believe daily. And the board should be warmed up when you run it by powering it up for a while.
More info:
https://www.ni.com/pdf/manuals/376812a.pdf
07-11-2022 03:16 AM
07-11-2022 06:38 PM
As per this link https://www.ni.com/en-us/shop/services/hardware/calibration-services/service-centers.html , there is a calibration centre in China